AMS 700 MAUDE Updates — May 2026

The final frontier. Deciding when, if and how.
lasthope2.0
Posts: 246
Joined: Sat Oct 11, 2025 1:23 pm

AMS 700 MAUDE Updates — May 2026

Postby lasthope2.0 » Mon Jun 22, 2026 10:19 pm

AMS 700 — May 2026 MAUDE Reported Device Age at Mechanical Malfunction

Summary Statistics
  • N: 76/154 with documented dates
  • Median: 3 years
  • Mean: 4 years
  • Max: 10 years

Percent Distribution

Code: Select all

≤1 yr     — 40%
1–4 yrs   — 18%
5–10 yrs  — 40%  (Similar to Titan)


Source: FDA MAUDE Database
Disclaimer: These are failure stats and not survival probability. Not professionally peer-reviewed and can have errors.

Summary of Events

1. INFECTION
  1. Infection is a recurring reason for explant — commonly scrotal infections with purulence, swelling, redness, pain, and pus.
  2. In many infection cases a temporary Tactra or malleable spacer was placed to preserve corporal space until the infection cleared.
  3. Infection types included recurrent UTIs, incision-site infections, bacterial/Staphylococcus infections, and chronic urethral-mucosa infection causing extrusion; some required full washout with antibiotics.
  4. Infection sometimes coexisted with other complications (e.g., a herniated and infected reservoir with inguinal abscess; post-implant pain later found to be scrotal infection). Classified as a known inherent risk per the IFU.

2. MALFUNCTION
  1. Fluid loss/leakage was the most common malfunction — holes, tears, pinholes, and peeling cylinder coating.
  2. Tubing failures — broken, kinked, cracked, or disconnected tubing, including a cluster of "crack in pump/cylinder connecting tube."
  3. Pump failures — won't inflate/deflate, sticky or dimpled pumps, micropunctures, faulty lock-out valves, and failed activation testing.
  4. Inflation/deflation problems — incomplete inflation, spontaneous/auto-deflation during intercourse, inflate-then-deflate; some no-leak cases were attributed to corporal stretch or scar capsule.
  5. Cylinder structural failures — bulges, aneurysmal change, buckling, layer dissociation, and fabric ingrowth.
  6. Lab analyses found wear at folds and kink-resistant tubing worn to filament; many holes were attributed to sharp-instrument damage during explant rather than a manufacturing defect.
  7. Reservoir malfunctions — empty reservoirs, tubing holes, and need for refill.

3. MALPOSITION
  1. Reservoir herniation — reservoir migrated out of the space of Retzius into the scrotum, groin, inguinal area, bladder, or ureter, causing pain.
  2. Pump migration — pump rotated, pulled up, or positioned too low or too deep posteriorly.
  3. Gland hypermobility ("floppy glans") — only the glans engorges while the shaft stays flaccid; some loss of glans sensation reported.
  4. Erosion / impending erosion — cylinders or pump eroding through the urethra or scrotum, including urethral extrusion.
  5. Crossover — distal or proximal crossover; sometimes attributed to original surgical misplacement (unintended use error).
  6. Cylinder migration — e.g., right cylinder migrating into the scrotum; cylinders folding over.
  7. Aneurysms — cylinder aneurysmal change (often left-sided) with bulging and loss of axial rigidity; some required planned corpora cavernosa reconstruction.

4. PATIENT DISSATISFACTION
  1. Size & rigidity — device too long/too large, sizing/measurement errors of roughly 3–4 cm, inadequate rigidity, Tenacio pump perceived as "too hard" or not fully inflating, and erections not staying stiff; often resolved by swapping a Tenacio pump for an MS pump.
  2. Cosmetic / placement concerns — pump placement or discomfort too low, reservoir bulge, cylinder herniation with dissatisfaction, and difficulty pumping/inflating.