2026 March Titan MAUDE Adverse Event Report Summary

Posted: Mon May 04, 2026 7:15 pm
by lasthope2.0
20 point summary of 129 event descriptions:
20-Point Summary

  1. Infection is the top clinical driver — 18 reports. Devices were explanted, often replaced with a Genesis malleable. Manufacturer narratives repeatedly state sterilization was validated and infection likely came from non-device sources.
  2. Tubing failure dominates mechanical complaints — 22 reports of fracture, break, tear, fray, or rupture, most commonly at the pump junction or near cylinders.
  3. Fluid loss / leaks — 13 reports. Patients noted loss of saline, hissing sounds, or inability to maintain inflation.
  4. Erosion / extrusion — 12 reports. Includes pump eroding through scrotum, tubing eroding through skin, cylinders extruding through urethral meatus, and lateral wall erosion.
  5. Pain, discomfort, or patient preference — 10 reports. Examples: pain near pump, reservoir bothering patient, or elective change from inflatable to malleable.
  6. Unspecified malfunction — 10 reports list only "unknown malfunction," "unspecified reason," or "device not working" with no further detail.
  7. Reservoir complications — 8 reports. Reservoir migration, herniation, extrusion, originally placed in wrong spot, or needing adjustment.
  8. Pump problems — 8 reports. Pump repositioned for being too high, pump stiff and won't pump, blocked activation/deactivation, and auto-inflation.
  9. Incorrect sizing — 7 reports. Undersized 18cm cylinders replaced with 20cm, 26cm replaced with 28cm, pump placed "too short," and surgeon dissatisfaction with tubing length.
  10. Failure to inflate — 6 reports. Cylinders would not fill, pump stays firm after 3 pumps, or "not cycling" with associated pain.
  11. Age-related revisions — 6 reports explicitly cite device age: 8, 13, and 15-year-old implants replaced electively with no defect found.
  12. Intraoperative damage — 6 reports. Cylinder material snapped while using pusher, eyelet punctured during implant, and one pump could not be prepped due to suspected fracture/air.
  13. Urethral injury / fistula — 5 reports. Includes urethral perforation during initial surgery, fistula with urine leaking into testicle, and staged repairs after cylinder extrusion.
  14. Cylinder pinhole / crack — 3 reports. One case confirmed by photo/video showing micro-hole in distal cylinder causing leak.
  15. Cylinder migration / malposition — 3 reports. Cylinders not reaching glans, or right cylinder sitting in wrong position.
  16. Cylinder aneurysm — 2 reports. One is a detailed case from Peyronie's disease with prior pericardial graft, requiring proximal extracapsular tunneling and graft-on-graft reinforcement.
  17. Kinked tubing — 1 report where tubing kink prevented inflation; tubing was trimmed, reconnected, pump reseated, device left in situ.
  18. Auto-inflation — 3 distinct reports of device re-inflating minutes after deflation, attributed to defective auto-inflate valve.
  19. Transgender anatomical incompatibility — 1 report notes current pump is too bulky with sharp corners, surgeons file them down; single cylinder's pointed tip causes partner pain and limits phalloplasty outcomes.
  20. Clinical literature — 2 entries
    • WMSM 2026 abstract: 66-year-old Peyronie's patient developed large cylinder aneurysm 2.5 years after grafted IPP; revision used proximal extracapsular tunneling + graft-on-graft Tutopatch to equalize corpora to 20cm and successfully re-implant.
    • NGS vs culture study (Henry et al., TAU 2025, DOI 10.21037/TAU-2025-57): In 9 men with early low-grade IPP infection, 6/9 failed antibiotics and needed explant within 7 days — shaft tenderness predicted early failure; study compares NGS to traditional cultures.

**UPDATE**
Of the mechanically malfunctioned devices, only 55 of them had documented implant and explant dates. Below is the percentage distribution of mechanical lifespan among these 55 failed samples:

≤1 year: 9%
2–5 years: 40%
6–10 years: 33%
11–15 years: 13%
16–20 years: 5%

Median lifespan: 6 years
Mean lifespan: 6.5 years
Maximum lifespan: 19 years

Re: 2026 March Titan MAUDE Adverse Event Report Summary

Posted: Wed May 06, 2026 4:32 am
by Fuckmachine700
I'm not sure what the point of this monthly news report is when no one seems to care. This site should be all about support, but regardless, I appreciate the hard work.

Re: 2026 March Titan MAUDE Adverse Event Report Summary

Posted: Wed May 06, 2026 9:15 am
by Thisworld
Fuckmachine700 wrote:I'm not sure what the point of this monthly news report is when no one seems to care. This site should be all about support, but regardless, I appreciate the hard work.

I do care and I think it's very valuable actually

Re: 2026 March Titan MAUDE Adverse Event Report Summary

Posted: Wed May 06, 2026 11:26 am
by lasthope2.0
Fuckmachine700 wrote:I'm not sure what the point of this monthly news report is when no one seems to care. This site should be all about support, but regardless, I appreciate the hard work.


Thank you for reading and your kind words. It’s about 129 real patients facing real problems. We’re often told complications occur in just 1% of cases. That 1% still represents real people whose lives have been impacted, not some lifeless statistics. I encourage them to come forward and share their problems and celebrations alike. I know this is nerdy and boring to read, but it still happened to a fellow bionic brother, and you and I may be one of them in the future. End of the day, it’s still OUR story!

Re: 2026 March Titan MAUDE Adverse Event Report Summary

Posted: Wed May 06, 2026 10:45 pm
by sambalamba
Fuckmachine700 wrote:I'm not sure what the point of this monthly news report is when no one seems to care. This site should be all about support, but regardless, I appreciate the hard work.


Maybe you don't care. But it is rubbish to think that nobody cares. I absolutely love it. Thanks to Lasthope for continuing to post these diligently. Keep up the good work.

Re: 2026 March Titan MAUDE Adverse Event Report Summary

Posted: Thu May 07, 2026 5:04 pm
by lasthope2.0
Of the mechanically malfunctioned devices, only 55 had documented implant and explant dates. Below is the percentage distribution of mechanical lifespan among these 55 failed samples:

≤1 year: 9%
2–5 years: 40%
6–10 years: 33%
11–15 years: 13%
16–20 years: 5%

Median lifespan: 6 years
Mean lifespan: 6.5 years
Maximum lifespan: 19 years

Re: 2026 March Titan MAUDE Adverse Event Report Summary

Posted: Thu May 07, 2026 5:49 pm
by sambalamba
lasthope2.0 wrote:Of the mechanically malfunctioned devices, only 55 had documented implant and explant dates. Below is the percentage distribution of mechanical lifespan among these 55 failed samples:

≤1 year: 9%
2–5 years: 40%
6–10 years: 33%
11–15 years: 13%
16–20 years: 5%

Median lifespan: 6 years
Mean lifespan: 6.5 years
Maximum lifespan: 19 years


Mean/Median lifespan in the failed devices never changes over the course of many of these reports. Indicative of lifespan for non failed devices? I think so.

Re: 2026 March Titan MAUDE Adverse Event Report Summary

Posted: Thu May 07, 2026 5:51 pm
by Kodixx
Yes I care very much also.

- Chuck
Thisworld wrote:
Fuckmachine700 wrote:I'm not sure what the point of this monthly news report is when no one seems to care. This site should be all about support, but regardless, I appreciate the hard work.

I do care and I think it's very valuable actually