Cycling issue

Posted: Fri Dec 20, 2019 6:25 pm
by WhiteCane
This freaking thing LOL... past two days, I’ve been doing my cycling as usual… it’s pumping up and getting erect but not exactly 100%… i’m up to 100 pumps before I said forget about it… I keep it at that and do the usual cycling routine… It doesn’t get rocksolid like I feel it should “the pump or the shaft of my penis“ do you think this could be a kinked tube? I’m 100% sure it’s not a leak because it’s definitely filling… Just not getting rocksolid in any capacity… Sort of frustrating… Any advice? Probably calling Kramer Monday…

Re: Cycling issue

Posted: Fri Dec 20, 2019 6:52 pm
by Gt1956
Silly question. Are you sure that you are getting a full pump each time? Investigate a little. You might be accidently getting less than full pumps in. Oh, any chance that you might be accidently pushing on the deflate button during the inflation cycle?

Re: Cycling issue

Posted: Fri Dec 20, 2019 7:00 pm
by tomas1
I guess worse case would be if there's not enough fluid in the system.

Re: Cycling issue

Posted: Fri Dec 20, 2019 7:16 pm
by WhiteCane
Pretty sure that I am but I will look into the full pumps thing… i’ve been till this point, it’s been no problem… How far should the release button stick out? maybe that’s stuck or something? Does this happen? can I fix it?

Re: Cycling issue

Posted: Fri Dec 20, 2019 8:09 pm
by Waynetho
WhiteCane wrote:This freaking thing LOL... past two days, I’ve been doing my cycling as usual… it’s pumping up and getting erect but not exactly 100%… i’m up to 100 pumps before I said forget about it… I keep it at that and do the usual cycling routine… It doesn’t get rocksolid like I feel it should “the pump or the shaft of my penis“ do you think this could be a kinked tube? I’m 100% sure it’s not a leak because it’s definitely filling… Just not getting rocksolid in any capacity… Sort of frustrating… Any advice? Probably calling Kramer Monday…


You have a Titan OTR or Touch, right? I recall someone here stating that the pump valve can get in a strange "half-way" state and pumping does nothing. I was going to say search here but maybe someone who's experienced the Titan Touch "micro-pump" issue (as I call it), can chime in and explain how they got theirs to pump properly. It seems I recall something about deflate then inflate again.

Re: Cycling issue

Posted: Fri Dec 20, 2019 8:45 pm
by WhiteCane
Oh I’m sorry… Still haven’t figured out the signature… Completely blind so it’s hard to navigate… I have an AMS device…

Re: Cycling issue

Posted: Fri Dec 20, 2019 9:24 pm
by oldbeek
How long have you had the implant? I had several glitches in the first months. To free a stuck release valve, Squeeze the valve block from end to end

Re: Cycling issue

Posted: Fri Dec 20, 2019 9:44 pm
by Waynetho
WhiteCane wrote:Oh I’m sorry… Still haven’t figured out the signature… Completely blind so it’s hard to navigate… I have an AMS device…


If the bulb doesn't appear to be filling up after you squeeze it, the release button may have been pressed while squeezing the bulb (the one thing everyone including Boston Scientific warns against doing).

What oldbeek said should resolve the issue if the deflate button was pressed while squeezing the pump. Find the SQUARE ENDS of the deflate block and squeeze. You should notice the bulb start to inflate again if done correctly. Once the bulb fills, press the deflate button, then try squeezing the bulb again - it should give a good "POP" and start inflating again.

If it consistently inflates but doesn't get completely inflated, the reservoir may be under filled or there may be a fluid leak somewhere. If the reset method doesn't fix it, I'd recommend doing nothing until you speak with Dr. Kramer or someone from his office.

Re: Cycling issue

Posted: Fri Dec 20, 2019 9:52 pm
by WhiteCane
Bingo! You guys are my heroes… needed to reset it… Thanks gents…