2025 October FDA Received AMS 700 Adverse Events

The final frontier. Deciding when, if and how.
lasthope2.0
Posts: 57
Joined: Sat Oct 11, 2025 1:23 pm

2025 October FDA Received AMS 700 Adverse Events

Postby lasthope2.0 » Thu Nov 20, 2025 9:53 pm

Attached - 2025 October FDA Received AMS 700 Adverse Events

Limited Mechanical Event Stats:
Only 41 out of 202 events had both event date and manufactured date
Assumption: 1 year of shelf life

Min: 2 months
Max: 12 yrs (Congratulations! 66 year old individual with LGX from South Korea. Event - Crack on pump connecting tube.)
Mean: 4 yrs
Median: 3 yrs

Sources:
-U.S. Food and Drug Administration Manufacturer and User Facility Device Experience (MAUDE) Database.
https://www.accessdata.fda.gov/scripts/ ... search.cfm

Disclaimer: MAUDE data is a passive surveillance tool with reporting limitations and should not be interpreted for causality.
You do not have the required permissions to view the files attached to this post.
Last edited by lasthope2.0 on Fri Nov 21, 2025 5:08 am, edited 2 times in total.
40, 2024 - Coloplast Genesis malleable 22cm-13mm with 1cm RTE. Subcoronal. Considering an IPP.

Thisworld
Posts: 283
Joined: Sun Sep 06, 2020 3:01 pm

Re: 2025 October FDA Received AMS 700 Adverse Events

Postby Thisworld » Fri Nov 21, 2025 2:56 am

Great work!
The average and median are calculated for mechanical events only , without considering infections and others biological issues, right?
Hard flaccid syndrome since 2019. Trying to get better with conservative treatments but an implant is on my radar

lasthope2.0
Posts: 57
Joined: Sat Oct 11, 2025 1:23 pm

Re: 2025 October FDA Received AMS 700 Adverse Events

Postby lasthope2.0 » Fri Nov 21, 2025 3:07 am

Thisworld wrote:Great work!
The average and median are calculated for mechanical events only , without considering infections and others biological issues, right?


Thank you, Thisworld. Yes, those are for mechanical events only. Many mechanical event reports had missing 'Date of Manufacture', so the observed mechanical events sample for the min/max/mean/median reduced to 41.

I have re-attached the Gemini high level summary PDF file after fixing an error. This is easier to digest compared to the tedious excel file. I've also included a high level failure mode summary and a note about Mayo Clinic's THALIA technique referenced for preventing reservoir migration.
You do not have the required permissions to view the files attached to this post.
40, 2024 - Coloplast Genesis malleable 22cm-13mm with 1cm RTE. Subcoronal. Considering an IPP.

User avatar
gregorbehr
Posts: 149
Joined: Wed May 26, 2010 5:39 am
Location: San Francisco, CA

Re: 2025 October FDA Received AMS 700 Adverse Events

Postby gregorbehr » Fri Nov 21, 2025 6:50 am

Thisworld wrote:Great work!
The average and median are calculated for mechanical events only , without considering infections and others biological issues, right?


It would be interesting what has been reported to them that would fall under the notion of biological issues.
51, San Francisco
Diabetic: used pills & injections.
2 TURP surgeries caused Peyronie's. 1st implant lost to infection, 2nd one is about 5 years old. But "floppy head" due to all the scar tissue, couldn't get tips all the way up.